Over time and especially if the EyeLink system has not been used for a while, the internal CMOS battery can discharge, which can reset BIOS settings to their defaults.
To correct the BIOS settings:
Over time and especially if the EyeLink system has not been used for a while, the internal CMOS battery can discharge, which can reset BIOS settings to their defaults.
To correct the BIOS settings:
- Please press F12 during startup - you should see a boot menu
- From here go into BIOS setup
- In the resulting screen under the System Configuration -> SATA Devices, make sure this is set to AHCI
- Save the settings and reboot